Magnetic resonance imaging (MRI) scans are often prohibited for patients with an
ID: 306367 • Letter: M
Question
Magnetic resonance imaging (MRI) scans are often prohibited for patients with an implanted device, such as a pacemaker. There are multiple reasons for this prohibition. One reason is the physical torque on the device due to the MRIs magnetic field. Suppose a paticent's pacemaker implant requires -1.6 mA of current to function. The leads of the devicetravel close together until they reach the heart and then split off to touch the top and bottom of the heart, such that size of the current loop is approximately half the cross-sectional area of the heart. Suppose the MRI machine uses a magnetic field of B 1.0T. Treating the heart as a box that is L = 12.0 cm long and W = 10.0 cm wide, calculate the magnitude of the maximum torque ? on the current loop.Explanation / Answer
Area of the heart
A=LW/2 =0.12*0.1/2
A=6*10-3 m2
Maximum torque acting on the current loop is
Tmax=NIAB =1*(1.6*10-3)*(6*10-3)*1
Tmax=9.6*10-6 N-m
Related Questions
Hire Me For All Your Tutoring Needs
Integrity-first tutoring: clear explanations, guidance, and feedback.
Drop an Email at
drjack9650@gmail.com