A gas chromatography column was operated under the following conditions: Column
ID: 517896 • Letter: A
Question
A gas chromatography column was operated under the following conditions:
Column dimension: 1.1 M length x 2.0 mm internal diameter Retention times: Nonretained carrier gas, tm = 0.30 min Compound A retention time, trA = 1.98 min Compound B retetntion time trb = 4.16 min Compound C retention time, trc = 7.93 min Peak widths at base : 0.18, 0.41, 0.82 min for A, B, and C respectively.
Calculate: i) the retention factor, k” for each compound ii) the selectivity factor, for each adjacent pair of compounds iii)
Explanation / Answer
Given,
tm = 0.3 min
trA = 1.98 min
trB = 4.16 min
trC = 7.93 min
i) Retention factor for A, kA = (trA - tm)/tm = (1.98 - 0.3)/0.3 = 5.6
Retention factor for B, kB = (trB- tm)/tm = (4.16-0.3)/0.3 = 12.86
Retention factor for C, kC = (trC- tm)/tm = (7.93-0.3)/0.3 = 25.43
ii) Selectivity factor for B and A , Alpha = kB / kA = 12.86/5.6 = 2.29
Selectivity factor for C and B , Alpha = kC / kB = 25.43/12.86 = 1.98