Below is the chromatogram of a standard solution of acenaphthene on a 50 mm L ti
ID: 520276 • Letter: B
Question
Below is the chromatogram of a standard solution of acenaphthene on a 50 mm L times 2.1 mm ID C18 column eluted isocratically with 50% acetonitrile in water at a flow rate of 0.2 ml/min. This standard solution is suggested by the vendor to check the efficiency of the column periodically. A Calculate the column efficiency (N) for this column, (use two equations and avg) B. What is the theoretical plate height (H) of this column under these conditions? C. A new column should have a column efficiency of -4400 for this standard under these conditions. Would you consider this column to be in good condition?Explanation / Answer
Ansa) Let us first write the condition given
column = acenapthalene 150mm L x 2.1mm ID C18
Phase = 50% acetonitrile in water
Flow rate = 0.2ml/min
Now from the chromatogram graph we can measure the retention time
Retention time t =760mm
and half width peak = 25mm
Now column efficiency N = 16 (t/w)2
= 16(760/25)2
= 14786
b) Now theoritical height H = Length of the column / number of theoritical plate
length of the column = 150mm
number of theoritical plates= 14786
H= 0.01014473mm
c) I would not consider this column to be in good condition as columns with high number of theoritical plates will have more efficiency and will produce narrower peaks .