Calibration of a glass electrode gave a reading of 128.9 mV with saturated potas

ID: 915101 • Letter: C

Question

Calibration of a glass electrode gave a reading of 128.9 mV with saturated potassium hydrogen tartrate buffer standard (pH = 3.557) and a reading of -98.8 mV with 0.008695 potassium dihydrogen phosphate, 0.0343 m disodium hydrogen phosphate buffer standard (pH = 7.400), both measured at 30°C.

(a) What is the observed slope (mV/pH unit) of the calibration curve?

(b) What is the theoretical slope at 30°C?

(c) What is , the electromotive efficiency to three significant figures?? Open the hint panel below for the definition of electromotive efficiency.

(d) What is the pH of an unknown that gives a reading of 24.0 mV with this electrode at 30°C?

Explanation / Answer