The following data for a mixture of hydrocarbons were obtained on a 25-m WCOT co

ID: 974420 • Letter: T

Question

The following data for a mixture of hydrocarbons were obtained on a 25-m WCOT column (0.25 mm i.d.) operated at 150EC. The liquid phase was SP-2100. The carrier gas flow rate was 30 mL/min

Compound

tr (min)

Peak Width (min)

Peak Area

Detector Response Factor

n-pentane

2.46

0.37

274000

0.630

n-heptane

4.47

0.41

761000

0.503

n-octane

6.53

0.43

319000

0.676

n-nonane

10.00

0.45

166000

0.618

a. Estimate the retention time for n-hexane.

b. Calculate the percentage of n-octane in this mixture.

c. Calculate the average number of theoretical plates for this column based on the data for n-pentane.

d. Calculate the resolution between n-heptane and n-octane. Is this resolution acceptable? Explain.

Compound

tr (min)

Peak Width (min)

Peak Area

Detector Response Factor

n-pentane

2.46

0.37

274000

0.630

n-heptane

4.47

0.41

761000

0.503

n-octane

6.53

0.43

319000

0.676

n-nonane

10.00

0.45

166000

0.618

Explanation / Answer

Answer

(a) around 3.5 minutes

(b)% of n-octane is 21.0%

(c)Theoretical plates for n-pentane is 707.3. where peak width is assumed which is equivalent to the peak width at 13.4 % of the peak height.

(d) 4.9 is resolution between n-heptane and n-octane. YES this much resolution is acceptable. Generally more than 1.0 resolution is acceptable in gas chromatography.