Calibration of a glass electrode gave a reading of 122.7 mV with 0.05 m potassiu
ID: 487183 • Letter: C
Question
Calibration of a glass electrode gave a reading of 122.7 mV with 0.05 m potassium dihydrogen citrate buffer standard (pH = 3.766) and a reading of -89.6 mV with 0.008695 potassium dihydrogen phosphate, 0.0343 m disodium hydrogen phosphate buffer standard (pH = 7.400), both measured at 30 degree C. (a) What is the observed slope (mV/pH unit) of the calibration curve? Slope_observed = (b) What is the theoretical slope at 30 degree C? Slope_theoretical = (c) What is beta, the electromotive efficiency to three significant figures?? Open the hint panel below for the definition of electromotive efficiency. beta = (d) What is the pH of an unknown that gives a reading of 13.8 mV with this electrode at 30 degree C? pH =Explanation / Answer
a)
Slope observed --> (mV2-mV1) / (pH2-pH1)
Slope observed --> (-89.6-122.7) / (7.4-3.766)
Slope = -58.420
b)
Slope, theoretical -->
Theorical slope = -0.1984 x T
T must be K = 30+273 = 303
so
theoretical --> -0.1984*303 = -60.115
c)
Beta = Slope observed/ theoretical
Beta =58.420/60.115 = 0.97180
d)
get pH for
13.8 mV
so
Slope observed --> (mV2-mV1) / (pH2-pH1)
-58.420 = (-89.6-13.8 ) / (7.4-pH)
solve for pH
pH = -( (-89.6-13.8 ) / -58.420 - 7.4) = 5.63